Joined: 1/4/2007 Last visit: 10/2/2024 Posts: 1143 Rating: (175)
|
Hello zamir209, How to generate such a Siwatool-file is descripted into the attached file, chapter 8. Important is, that you “receive all data” at first, before you are saving the file. Can you connectone load cell directly at the Siwarex U module? So we can be sure, that the Siwarex U module is ok. I guess, that there is something wrong with the wiring between the Siwarex U module and the load cell. Which load cell(s) are you using? How is the correct wiring of that load cell(s)? Please send to me the technical data sheet of the load cell(s). Best regards Siwarexi AttachmentSIWAREX_U_Quick Guide_V3.1.pdf (191 Downloads) |
5/10/2010 3:49 PM | |
Joined: 1/4/2007 Last visit: 10/2/2024 Posts: 1143 Rating: (175)
|
|