6/19/2009 12:17 PM | |
Joined: 1/4/2007 Last visit: 8/26/2025 Posts: 1190 Rating:
Dear Abyrne, To your questions: How did you calculate the first 3 results to give a Kg result? My answer: I used the formula “formula.jpg”, which you can see into my last reply. Your question: Am i correct in saying the formula gives an answer as a % of error? My answer: No, the result is in kg. This kg-value is valid from an unloaded load cell up to the nominal load of a load cell. Attached you can also have my Excel-sheet for the calculation. Your question: I also would like to know how drift is calculated with respect to time... i.e. what % of drift is there over a 1 or 2 year period? My answer: The drift is valid for the life cycle of the load cell. If the load cell has a higher drift, the load cell is damaged, you have mechanical problems (like bypass forces, frictions, ...) or the wires of the load cell cable are not connected correctly. Best regards Siwarexi Attachmenterror_limit.zip (320 Downloads) |
