12/13/2024 1:25 PM | |
Joined: 1/4/2007 Last visit: 8/26/2025 Posts: 1190 Rating:
|
Hello Agustin Ordoñez, Please execute the command “Set calibration point 1 (61)” in Siwatool once again, when your calibration weight of 195 grams is on the scale. When the error occurred, then: DO NOTHING AND create a Siwatool-file. Attached you can find a pdf-file how to create a Siwatool file. Post your Siwatool file here in the Forum. I would like to check your Siwatool file to find the issue. Best regards Siwarexi
Attachmentcreate_a_SIWATOOL-file.pdf (236 Downloads) |
12/13/2024 2:32 PM | |
Joined: 8/30/2023 Last visit: 8/26/2025 Posts: 24 Rating:
|
Hi Siwarexi, thanks for your response. Here is the file in a txt format because forum wouldn't let me upload it with wp321 file extension I noticed I get the following error: "7007 Calibrationweights / Calibrationdigits not plausible" AttachmentCalibrationError.WP321.txt (198 Downloads) |
12/13/2024 3:21 PM | |
Joined: 8/30/2023 Last visit: 8/26/2025 Posts: 24 Rating:
|
Thanks for your response, I've tried flipping the cells upside down and they now increase the mv/v value when pressed down, but I still get the same error I'll attach the corresponding file AttachmentCalibrationError2.WP321.txt (220 Downloads) |
12/17/2024 9:36 AM | |
Joined: 1/4/2007 Last visit: 8/26/2025 Posts: 1190 Rating:
|
Hello Agustin Ordoñez, Unfortunately, in your Siwatool-file, in DR3 the “Calibration digits 0” has the value 0, and the “Calibration digits 1” has the value 2000000. These values are only factory settings! Therefore, I can’t see the issue! Please execute the command “Set calibration point 1 (61)” in Siwatool once again, when your calibration weight of 195 grams is on the scale. When the error occurred, then: DO NOTHING (= let the calibration weight on the scale) AND create a Siwatool-file. Attached you can find a pdf-file how to create a Siwatool file. Post your Siwatool file here in the Forum. I would like to check your Siwatool file to find the issue. Best regards Siwarexi
Attachmentcreate_a_SIWATOOL-file.pdf (215 Downloads) |
12/19/2024 1:10 PM | |
Joined: 8/30/2023 Last visit: 8/26/2025 Posts: 24 Rating:
|
Hi, I was able to diagnose the issue with those cells, cell started at -0.3mv/v and can't calibrate with less than 0. Just need to add some more weight! Now I am having issue with the real scale (3 30tons cells). I am trying to calibrate with my body weight right now and I get the same issue. Here is the siwatool file AttachmentCalibrationError3.WP321.txt (216 Downloads) |