Hi. I am having a similar problem in a Siemens Healthineers old MRI, The patient table occasionally refuses to start going down and the Micromaster shows the F002 error until I reboot the system and the loadware is set again. it drives a spindle that will carry down the patient table. I am suspecting the spindle not smooth due to poor greasing history. Any suggestions? Thanks! Hassan ------------------------------------------------------------------------------------------ |
Last edited by: Moderator_Lan at: 02/17/2025 10:34:50New subject after splitting |
|
2/18/2025 2:44 PM | |
Joined: 3/27/2008 Last visit: 4/8/2025 Posts: 1190 Rating:
|
The F0002, is over Voltage, this can be initiated by either a high supply Voltage, or the load is being stopped too quickly, i.e. the ramp down time is too short and the drive cannot absorb the inertia during the ramp down. Ramp down time is set by P1121, try making this time longer to see if the F0002 goes away? Tractor |