1/17/2016 4:43 PM | |
Joined: 9/27/2006 Last visit: 5/19/2024 Posts: 12257 Rating: (2668) |
Hello M.Afroukh; The part number that you mention is not correct for an Ultramat 6 unit:
ULTRAMAT 6 and OXYMAT 6
Analyzers for IR-absorbing gases and oxygen
7MB2011, 7MB2021, 7MB2023, 7MB2024, 7MB2111, 7MB2112, 7MB2121, 7MB2123, 7MB2124
7MB2017, 7MB2027, 7MB2028, 7MB2026, 7MB2117, 7MB2118, 7MB2127, 7MB2128, 7MB2126
https://support.industry.siemens.com/cs/ww/en/view/7618021 Chapter 7 of the manual shows a spareparts list. Check if the keyboard you are looking for is mentioned. If the part number you mention is not available in the manual, I would suggest you open the Services link above (at the right) and look for the Spare parts department of Siemens. You can ask them directly for specific part you need replaced (it would probably be best to have the part number of the original Ultramat unit on hand when you request the spare keyboard unit, as they might have issues with hardware/software versions). Hope this helps, Daniel Chartier |
Last edited by: dchartier at: 1/17/2016 5:36:27 PMLast edited by: dchartier at: 1/17/2016 6:17:04 PM |
|
1/28/2016 8:34 AM | |
Joined: 11/20/2007 Last visit: 6/16/2023 Posts: 25 Rating: (3)
|
Hello M.Afroukh, Note: It is only possible to replace the display and not the keypad. |
This contribution was helpful to1 thankful Users |