5/23/2016 10:16 AM | |
Joined: 4/26/2016 Last visit: 6/1/2022 Posts: 22 Rating: (1) |
Hello all, Anybody can help me how to calculate index number if I use DN 2600 (104 in)? Regards, |
6/29/2016 7:50 AM | |
Joined: 3/28/2010 Last visit: 12/10/2024 Posts: 1097 Rating: (216) |
I attended a Siemens flow seminar last week and the presenter mentioned that Si-Ware software could be used to determine the transducer model and to calculate the LTN distance-between-transducer-faces value. I had him show me where and how it's done (Utilities tab > Liquid Tool) The attached pdf document shows a screen shot from Si-Ware for a calculation for 102 inch carbon steel pipe, water at flow rates from 1 to 20 feet/sec. The resulting LTN distance is 117.2", a little less than 3m. You can download Si-Ware software and run the calcs yourself in SI units. AttachmentTransducer and LTN calc for 102 inch OD pipe, 1-20fps flow rate.pdf (173 Downloads) |
This contribution was helpful to1 thankful Users |
8/3/2016 3:27 AM | |
Joined: 4/26/2016 Last visit: 6/1/2022 Posts: 22 Rating: (1) |
Thank you so much for danw and Fl0w for your information, it's very helpful, thanks guys. |