I bought MAG 6000 I, model code 7ME6930-2BA52-1JA0, with pulse output but my agent said the pulse output cannot give totalized value to PLC. Is it real for that? Is it only using bus communication to send totalized value to PLC? I read manual, there is only totalizer 2 cannot read from pulse output. Is it correct?
---------------------------------------------------------------------------------------------------- |
Last edited by: Min_Moderator at: 7/28/2015 3:52:37 PMNew subject after splitting |
|
7/30/2015 2:49 AM | |
Posts: 6 Rating: (1) |
I don't decode or decrypt long model numbers to figure out functionality, but if your Mag 8000 has a pulse output, then my Socratic observations are: 1) Why do you suppose Siemens prints manuals for its products? Like A5E03071515-007 dated 05/2014? 2) Why do you suppose a Mag8000 manual can be downloaded from this link? 3) What does the Pulse Output configuration (section 6.5, pages 51-52) say the pulse output can be configured as? 4) Why is there a parameter for pulse output configuration called Amount per pulse A with units like m3? 5) What does the expanded explanation on page 93 (pdf) tell you? 6) What is the result of counting pulses where each pulse represents a specific volume? |