If your customer can tolerate ±2% error, then Siemens has documentation from some European agency that certifies the Mag 8000 as having error less than ±2% error for custody transfer applications with zero-upstream and/or Zero downstream straight run. There's a pdf brochure stating this at this URL: https://tinyurl.com/https-www-industry-usa-sieme The statement of interest is "Based on its demonstrated ability to perform in non-ideal conditions, the MAG 5100W and MAG 8000 have received approval for 2% accuracy custody transfer (MI-001) and non custody transfer applications for installation with zero diameter upstream pipe and zero diameter downstream straight run pipe from the sensor." At one time there were photos floating around of the flow test stand showing offset elbows and tees on the inlet or outlet of the magmeter under test. I was astounded because for years we'd be warned of the dire consequences of not having sufficient upstream/downstream straight run, but without any citation of what the magnitude of the error would be. And determining the error has never been a do-it-yourself project - it requires a flow stand which is a rare beast to have access to. Dan |