7/4/2018 4:40 PM | |
Posts: 206 Rating: (22) |
Hi Lèvy, if you want to display two SecureDisplays on the same panel, you have to edit this in the "DisplayCali.xml" file of the SecureDisplay. There you have to duplicate the structure "DisplayEntry" and edit the IP address to the second Siwarex WP251. The "Ready for use" https://support.industry.siemens.com/cs/ww/en/view/109481714 section "„Ready-for-use“ for stand-alone operation without SIMATIC CPU" is prepared for up to 8 Siwarex WP251. The only thing you have to do is to edit the IP addresses in the "Connections" of WinCC and to set the variable "Mux" to 0 for the first Siwarex WP251 or to 1 for the second. If you have done this the display will access the appropriate Siwarex WP251. There in picture 6.2.0 you have a button for zeroing and tareing. How to set the "Mux" variable you can see in the "Ready for use" of section "„Ready-for-use“ for operation with SIMATIC CPU and eight SIWAREX WP251". Regards ABenno |