3/26/2025 2:06 PM | |
Joined: 3/28/2010 Last visit: 9/12/2025 Posts: 1172 Rating:
|
download the Mag5000 manual https://support.industry.siemens.com/cs/de/en/view/85671888 Read the sections on "Indicates flow with no flow in pipe" and "Unstable flow signal" on page 63 pdf. |
Last edited by: Moderator_Lan at: 03/26/2025 15:13:04link optimized |
|
This contribution was helpful to
1 thankful Users |
3/27/2025 5:50 AM | |
Joined: 11/7/2024 Last visit: 7/29/2025 Posts: 2 Rating:
|
Thank you, Danw, for your response! Just to clarify, the issue I'm encountering is that the value shown on the flowmeter display is fluctuating. After reading the manual, I believe the "Unstable flow signal" refers to the signal output rather than the display itself. |
3/27/2025 2:11 PM | |
Joined: 3/28/2010 Last visit: 9/12/2025 Posts: 1172 Rating:
|
The analog output mimics the display, The two have effectively the same values simultaneously. |
Follow us on