Hello Dinh; If you believe that a titanium electrode is required because of the corrosive medium, then you could look at the Mag 3100, or Mag 3100HT for higher temperature range. Look at the Sitrans F flowmeter catalog for more details: http://www.automation.siemens.com/w1/efiles/instrumentation/catalogs/fi01/fi01_en_unit/FI01_2010_en_kap04.pdf On the other hand, only the Mag 1100F is considered food-grade in electromagnetic sensors, so that might be a problem too. Maybe you could consider a clamp-on ultrasonic flowmeter, such as the FUS 1010 standard, which required no contact with the medium, can be ordered with a high-temperature transducer, and provides accuracy to ± 0.5% ... 1.0% of flow, for velocities greater than 0.3 m/s (1 ft/s) on a steel pipe? In any case, whatever your selection, the only one who can give you commercial information, such as lead time, delivery and pricing would be your local Siemens representative. He can contact the Sitrans sensorsdivision plants to get all this info for you. Hope this helps, Daniel Chartier |