12/15/2011 7:31 AM | |
Posts: 3609 Rating: (29) |
|
12/16/2011 2:58 AM | |
Joined: 3/28/2010 Last visit: 9/18/2024 Posts: 1056 Rating: (213) |
You might want to review the document "MODBUS RTU RS 485 add-on module for USM II transmitters". It appears to be document # FDK:521H1183. It covers the use of the MODBUS RTU add-on module, product code number FDK:085U0234, that can be used in the Siemens Flow Instruments USM II family of transmitters, which includes MAG 6000. The add-on module makes the Mag 6000 into a Modbus RTU slave on a 2 wire RS-485 network. A Modbus master sends a 03 read instruction/command/function code (read holding register) requesting data from registers 43002 and 43003. The slave returns the contents of those registers, a floating point (real) value that is the absolute volume in m3/Sec. AttachmentModbus RTU over RS485 Mag 6000 manual SFIDKPS023D302.pdf (425 Downloads) |
Last edited by: danw at: 12/16/2011 2:58 AM |
|
Follow us on