1/25/2011 12:55 PM | |
Posts: 88 Rating: (2) |
Thank you guys! another question about Analog inputs of em 231 i simulate a 4-20mA and with the Gain potentiometer i try to calibrate it to show me 6400-32000(as 4-20mA) the problem every time i change one of the limits the second changes in the same direction too, is there way to do so? and my configuration dip switches configured 1,2-ON 3-6-off. Best Regards, Greg. Splitted from S7-224XP on board Analog IO. |
Last edited by: O_Moderator at: 1/26/2011 10:36 AMnew subject after splitting |
|
1/25/2011 1:45 PM | |
Joined: 4/22/2010 Last visit: 10/17/2024 Posts: 5697 Rating: (715) |
That's what the gain does, it should be adjusted with full signal (20mA) so that the reading is 32000. To adjust the zero point you need a offset and that is not on your module only on the 235. maybe a usefull link: http://support.automation.siemens.com/WW/view/en/22148598 |
Last edited by: Marcjan at: 1/25/2011 1:56 PMLast edited by: Marcjan at: 1/25/2011 1:46 PMProblem solved? |
|
1/25/2011 2:19 PM | |
Posts: 88 Rating: (2) |
Thanks! |