6/1/2011 12:35 AM | |
Posts: 95 Rating: (0) |
Hi, H-H and Ayman Elotaify, I fully understand the number 27648 now and Iget that10885.04 is 27648 divided by 2.54 cm/in. I stillcan't figure outthe number 346.6570, but I think it won't be a big problembecause I can create a new number based on the 27648 theory when I calibrate the tank. Thank you guys. Best reagrds |
6/1/2011 8:48 PM | |
Posts: 95 Rating: (0) |
Thanks Ayman. There are two functions for the same pressure transmitters which are 4-20mA, 0-100 inch water; The equation of Pressure to Volume is like this: V = (AI / 3466.57 * Range * r * r ) / 100000 + C V: Volume, unit: liter or m^3 The equation of Pressure to Height is like this: H = AI * Range / 10885.04 = AI * range / (27648/2.54) H: height in meters; Best reagrds |
6/1/2011 11:47 PM | |
Joined: 7/2/2008 Last visit: 8/26/2024 Posts: 928 Rating: (307) |
Hello AlecTek,
V = π*h*r2 V= 3.14 * (AI/10885.04) * r2 V= (AI / 3446.57) * r2 I guess you know now where the 3446.57 comes from. Now h (AI/10885.04) is in meter and if the radius is in meter as well you get a volume in m3. To convert it to liters just multiply by 1000. Is it still not clear? |
This contribution was helpful to2 thankful Users |
6/2/2011 1:16 AM | |
Posts: 95 Rating: (0) |
Oh I got it. I am so stupid! |
Follow us on