7/1/2016 3:38 PM | |
Joined: 1/4/2007 Last visit: 12/8/2023 Posts: 1071 Rating:
|
Hello rQx, “Is maybe the reason that they only did the "correction factor" instead of "span calibration" because we have the metal hinge?” NO! A calibration must be done - in every case - by executing the two steps: !!! If you can’t see the correct weight value, the calibration was done not correctly!!! A correction factor of 2.3 is quite high. The other belt scale with the correction factor of 13.5 makes absolutely no sense. A wrong belt length means that the time for the calibration is not correct. “With our construction would it be enough to only do the zero calibration, and then the span calibration with test weight.” “Or do we have to use the correction factor also because we have the metal hinge?” Do at first a correct calibration.
“And how important is the "angle of the belt" if the angle is always the same?”
|